WADO retrieve from Osiris not possible

Hallo!

I got yet another problem/question:

When I set the retrieve mode to WADO in Osirix (Horos) or MedFilm using the iPad, the images cannot be received. It works when using C-MOVE. The dicom-web plugin seems to run fine from the logs.

When initiating a retrieve Orthanc log stops here:

I0515 12:26:47.106884 CommandDispatcher.cpp:491] Association Received from AET HOROS on IP 10.0.1.3
I0515 12:26:47.107568 CommandDispatcher.cpp:689] Association Acknowledged (Max Send PDV: 16372)
I0515 12:26:47.205007 main.cpp:120] No limit on the number of C-FIND results at the Patient, Study and Series levels
I0515 12:26:47.205326 main.cpp:130] No limit on the number of C-FIND results at the Instance level
I0515 12:26:47.205728 OrthancFindRequestHandler.cpp:566] DICOM C-Find request at level: Instance
I0515 12:26:47.205928 OrthancFindRequestHandler.cpp:572] (0008,0005) SpecificCharacterSet = ISO_IR 100
I0515 12:26:47.206123 OrthancFindRequestHandler.cpp:572] (0008,0018) SOPInstanceUID =
I0515 12:26:47.206283 OrthancFindRequestHandler.cpp:572] (0008,0052) QueryRetrieveLevel = IMAGE
I0515 12:26:47.206471 OrthancFindRequestHandler.cpp:572] (0020,000d) StudyInstanceUID = 1.2.40.0.23.6616.1.2332904
I0515 12:26:47.206675 OrthancFindRequestHandler.cpp:572] (0020,000e) SeriesInstanceUID =

Horos gives me this error:

DICOM Network Failure (query)
SCU Failed 0006:0207 DIMSE No data available (timeout in non-blocking mode). Try to increase the timeout parameter in the Listener Preferences.

THANK YOU for any help!
regards

Hello,

I don’t personally have access to an Apple computer, so I cannot help you directly. In particular, I’m unsure of what “WADO” means in the context of OsiriX/Horos (Is it C-GET, WADO-RS, or WADO-URI?).

Please could you post the full log of Orhanc, with the “–verbose” flag? This may give hints.

Sébastien-

Sorry, I meant with the “–trace” flag.

Thanks you Sébastien for the fast response.

OK, I found out that it works with just a few patients in the storage, but not with several hundred.
When I have lots of patients, Orthanc log (with --trace) gives me nearly endless lines of:

T0515 14:19:39.138177 Statement.cpp:144] SQLite::Statement::Step SELECT internalId FROM Resources WHERE resourceType=?

followed by:

T0515 14:19:39.147308 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?

and then

T0515 14:19:39.182038 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.182267 Statement.cpp:144] SQLite::Statement::Step SELECT uuid, uncompressedSize, compressionType, compressedSize, uncompressedMD5, compressedMD5 FROM AttachedFiles WHERE id=? AND fileType=?

finally the WADO retrieve starts:

I0515 14:19:39.326519 OrthancFindRequestHandler.cpp:666] Number of matching resources: 2
I0515 14:19:39.339214 CommandDispatcher.cpp:860] DUL Peer Requested Release
I0515 14:19:39.339448 CommandDispatcher.cpp:867] Association Release
T0515 14:19:39.467626 MongooseServer.cpp:604] HTTP header: [host]: [10.0.1.2:8042]
T0515 14:19:39.467974 MongooseServer.cpp:604] HTTP header: [accept]: [*/*]
T0515 14:19:39.467967 MongooseServer.cpp:604] HTTP header: [host]: [10.0.1.2:8042]
T0515 14:19:39.468473 MongooseServer.cpp:604] HTTP header: [accept]: [*/*]
T0515 14:19:39.468464 MongooseServer.cpp:604] HTTP header: [accept-language]: [de-de]
T0515 14:19:39.468855 MongooseServer.cpp:604] HTTP header: [accept-language]: [de-de]
T0515 14:19:39.468865 MongooseServer.cpp:604] HTTP header: [connection]: [keep-alive]
T0515 14:19:39.469341 MongooseServer.cpp:604] HTTP header: [connection]: [keep-alive]
T0515 14:19:39.469348 MongooseServer.cpp:604] HTTP header: [accept-encoding]: [gzip, deflate]
T0515 14:19:39.469751 MongooseServer.cpp:604] HTTP header: [accept-encoding]: [gzip, deflate]
T0515 14:19:39.469757 MongooseServer.cpp:604] HTTP header: [user-agent]: [Horos/20160210 CFNetwork/760.5.1 Darwin/15.5.0 (x86_64)]
I0515 14:19:39.470335 MongooseServer.cpp:723] GET /wado
T0515 14:19:39.470331 MongooseServer.cpp:604] HTTP header: [user-agent]: [Horos/20160210 CFNetwork/760.5.1 Darwin/15.5.0 (x86_64)]
I0515 14:19:39.470727 OrthancPlugins.cpp:560] Delegating HTTP request to plugin for URI: /wado
I0515 14:19:39.470968 MongooseServer.cpp:723] GET /wado
T0515 14:19:39.471101 OrthancPlugins.cpp:1769] Calling service 3008 from plugin ./libOrthancDicomWeb.dylib
I0515 14:19:39.471109 OrthancPlugins.cpp:560] Delegating HTTP request to plugin for URI: /wado
T0515 14:19:39.471564 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.472058 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.472462 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.472671 OrthancPlugins.cpp:1769] Calling service 3007 from plugin ./libOrthancDicomWeb.dylib
T0515 14:19:39.472892 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.473322 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.473727 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.473974 OrthancPlugins.cpp:1769] Calling service 3001 from plugin ./libOrthancDicomWeb.dylib
I0515 14:19:39.474161 OrthancPlugins.cpp:952] Plugin making REST GET call on URI /instances/7988b8d4-5b97e8c9-440a0674-9108848a-4353d6ad/series (built-in API)
T0515 14:19:39.474420 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.474654 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:39.474883 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.475096 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.475328 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:39.475541 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.475774 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:39.475983 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:39.476233 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.476424 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.476615 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.476795 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.476965 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.477136 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.477329 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.477511 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.477679 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.477846 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.478012 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.478178 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.478349 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.478539 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.478784 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.478945 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.479166 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.479458 OrthancPlugins.cpp:1769] Calling service 3006 from plugin ./libOrthancDicomWeb.dylib
T0515 14:19:39.479703 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.480079 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.480329 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.480546 OrthancPlugins.cpp:1769] Calling service 3001 from plugin ./libOrthancDicomWeb.dylib
I0515 14:19:39.480684 OrthancPlugins.cpp:952] Plugin making REST GET call on URI /instances/7988b8d4-5b97e8c9-440a0674-9108848a-4353d6ad/study (built-in API)
T0515 14:19:39.480903 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.481110 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:39.481299 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.481519 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.481687 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:39.481879 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.482071 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.482274 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:39.482497 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.482722 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:39.482945 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:39.483172 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:39.483445 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.483639 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.483781 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.483964 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.484098 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.484263 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.484423 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.484602 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.484774 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.484949 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.485139 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.485270 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.485429 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.485672 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.485824 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.486001 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.486311 OrthancPlugins.cpp:1769] Calling service 3001 from plugin ./libOrthancDicomWeb.dylib
I0515 14:19:39.486475 OrthancPlugins.cpp:952] Plugin making REST GET call on URI /instances/7988b8d4-5b97e8c9-440a0674-9108848a-4353d6ad/file (built-in API)
T0515 14:19:39.486717 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.486908 Statement.cpp:144] SQLite::Statement::Step SELECT uuid, uncompressedSize, compressionType, compressedSize, uncompressedMD5, compressedMD5 FROM AttachedFiles WHERE id=? AND fileType=?
T0515 14:19:39.492484 OrthancPlugins.cpp:1769] Calling service 2000 from plugin ./libOrthancDicomWeb.dylib
T0515 14:19:39.518218 Connection.cpp:388] SQLite::Connection::FlushToDisk
T0515 14:19:39.589650 HttpOutput.cpp:368] Compressing a HTTP answer using gzip
T0515 14:19:39.686035 OrthancPlugins.cpp:1769] Calling service 3008 from plugin ./libOrthancDicomWeb.dylib
T0515 14:19:39.686676 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.687349 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.687754 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.688032 OrthancPlugins.cpp:1769] Calling service 3007 from plugin ./libOrthancDicomWeb.dylib
T0515 14:19:39.688401 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.688785 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.689123 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.689542 OrthancPlugins.cpp:1769] Calling service 3001 from plugin ./libOrthancDicomWeb.dylib
I0515 14:19:39.689755 OrthancPlugins.cpp:952] Plugin making REST GET call on URI /instances/e1bac10b-8aa26b3c-725abc59-0b92b915-abee8b64/series (built-in API)
T0515 14:19:39.690058 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.690396 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:39.690651 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.690909 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.691137 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:39.691369 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.691595 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:39.691899 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:39.692175 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.692402 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.692648 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.692987 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.693172 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.693361 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.693576 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.693993 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.694126 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.694312 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.694505 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.694632 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.694798 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.694967 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.695176 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.695454 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.695601 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.695919 OrthancPlugins.cpp:1769] Calling service 3006 from plugin ./libOrthancDicomWeb.dylib
T0515 14:19:39.696180 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.696503 Statement.cpp:144] SQLite::Statement::Step SELECT d.id FROM DicomIdentifiers AS d, Resources AS r WHERE d.id = r.internalId AND r.resourceType=? AND d.tagGroup=? AND d.tagElement=? AND d.value=?
T0515 14:19:39.696875 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.697077 OrthancPlugins.cpp:1769] Calling service 3001 from plugin ./libOrthancDicomWeb.dylib
I0515 14:19:39.697237 OrthancPlugins.cpp:952] Plugin making REST GET call on URI /instances/e1bac10b-8aa26b3c-725abc59-0b92b915-abee8b64/study (built-in API)
T0515 14:19:39.697506 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.697703 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:39.697887 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.698103 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.698341 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:39.698548 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.698715 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.698922 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:39.699145 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:39.699306 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:39.699549 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:39.699781 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:39.700027 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.700208 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.700376 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.700542 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.700711 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.700873 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.701040 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.701205 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.701370 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.701546 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.701693 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.701803 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.701920 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:39.702127 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.702276 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.702451 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:39.702767 OrthancPlugins.cpp:1769] Calling service 3001 from plugin ./libOrthancDicomWeb.dylib
I0515 14:19:39.702934 OrthancPlugins.cpp:952] Plugin making REST GET call on URI /instances/e1bac10b-8aa26b3c-725abc59-0b92b915-abee8b64/file (built-in API)
T0515 14:19:39.703167 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:39.703359 Statement.cpp:144] SQLite::Statement::Step SELECT uuid, uncompressedSize, compressionType, compressedSize, uncompressedMD5, compressedMD5 FROM AttachedFiles WHERE id=? AND fileType=?
T0515 14:19:39.709021 OrthancPlugins.cpp:1769] Calling service 2000 from plugin ./libOrthancDicomWeb.dylib
T0515 14:19:39.776787 HttpOutput.cpp:368] Compressing a HTTP answer using gzip
T0515 14:19:49.556507 Connection.cpp:388] SQLite::Connection::FlushToDisk
T0515 14:19:50.240287 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM Resources WHERE internalId=?
T0515 14:19:50.240781 Statement.cpp:135] SQLite::Statement::Run INSERT INTO Changes VALUES(NULL, ?, ?, ?, ?)
T0515 14:19:50.345009 ServerIndex.cpp:181] Change related to resource f3496489-59315f3e-2643b923-fc5016fb-0ec7fa59 of type Series: StableSeries
T0515 14:19:50.345279 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM Resources WHERE internalId=?
T0515 14:19:50.345499 Statement.cpp:135] SQLite::Statement::Run INSERT INTO Changes VALUES(NULL, ?, ?, ?, ?)
T0515 14:19:50.345730 ServerIndex.cpp:181] Change related to resource b653ccbd-796885b4-0594b4fa-7fef2c50-18de563e of type Study: StableStudy
T0515 14:19:50.345958 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM Resources WHERE internalId=?
T0515 14:19:50.346153 Statement.cpp:135] SQLite::Statement::Run INSERT INTO Changes VALUES(NULL, ?, ?, ?, ?)
T0515 14:19:50.353750 ServerIndex.cpp:181] Change related to resource 09aa615f-4d57a333-e310562d-61b644c8-735308e5 of type Patient: StablePatient
T0515 14:19:50.353953 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:50.354180 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:50.354351 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:50.354535 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.354840 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.355129 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.355441 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.355750 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.356060 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.356318 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.356574 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.356829 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.357088 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.357432 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.357701 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.357994 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.358266 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.358530 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.358790 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.359051 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.359312 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.359573 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.359835 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.360097 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.360358 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.360694 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.360954 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.361221 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.361480 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.361771 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.362072 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.362331 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.362597 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.362929 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.363185 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.363469 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.363779 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.364098 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.364360 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.364619 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.364911 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.365306 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.365580 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.365830 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.366100 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.366355 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.366603 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.366852 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.367108 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.367416 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.367712 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.367955 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.368211 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.368459 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.368619 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.368849 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.369104 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.369352 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.369603 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.369853 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.370100 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.370349 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.370600 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.370860 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.371147 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.371391 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.371619 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.371862 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.372104 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.372342 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.372522 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.372735 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.373006 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.373227 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.373495 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.373765 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.374047 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.374267 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.374478 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.374743 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.374937 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.375161 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.375430 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.375720 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.375988 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.376266 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.376528 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.376720 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.376982 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.377265 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.377557 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.377821 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.378049 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.378280 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.378652 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.378858 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.379084 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.379270 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.379492 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.379666 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.379850 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.380032 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.380253 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.380437 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.380667 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.380827 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.380983 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.381195 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.381400 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.381585 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.381818 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:50.382028 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:19:50.382193 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:19:50.382348 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.382597 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:50.382803 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:19:50.382990 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:19:50.383178 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.383444 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.383694 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.383954 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.384125 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.384283 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.384441 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.384597 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.384752 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.384906 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.385061 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.385217 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.385370 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.385526 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.385683 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.385837 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.386049 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.386237 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.386422 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.386613 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:50.386820 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:19:50.386982 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:19:50.387139 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.387347 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:50.387553 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.387814 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:19:50.388176 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.388352 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.388508 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.388664 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.388820 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:19:50.388994 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.389211 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.389403 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:50.389599 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:19:50.389807 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:19:50.389971 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:19:50.390125 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:19:59.585401 Connection.cpp:388] SQLite::Connection::FlushToDisk
T0515 14:20:09.785306 Connection.cpp:388] SQLite::Connection::FlushToDisk
T0515 14:20:11.419541 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM Resources WHERE internalId=?
T0515 14:20:11.420039 Statement.cpp:135] SQLite::Statement::Run INSERT INTO Changes VALUES(NULL, ?, ?, ?, ?)
T0515 14:20:11.520577 ServerIndex.cpp:181] Change related to resource b0ad9234-fd3eba9c-e6b4ad68-fef279dc-ac2d3969 of type Series: StableSeries
T0515 14:20:11.521009 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:20:11.521378 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:20:11.521705 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:20:11.522029 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:20:11.522433 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:20:11.522827 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:11.523150 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:11.523485 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.523771 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.523965 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.524205 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.524468 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.524650 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.524878 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.525107 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.525378 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.525559 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.525763 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.526003 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:11.526291 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:11.526477 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:11.526748 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:11.527002 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:20:11.527202 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:20:11.527402 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:20:11.527593 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.525596 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM Resources WHERE internalId=?
T0515 14:20:12.525967 Statement.cpp:135] SQLite::Statement::Run INSERT INTO Changes VALUES(NULL, ?, ?, ?, ?)
T0515 14:20:12.526342 ServerIndex.cpp:181] Change related to resource 3968e539-f28817bc-afe86ae1-7058a2c0-f1188255 of type Series: StableSeries
T0515 14:20:12.526789 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM Resources WHERE internalId=?
T0515 14:20:12.527167 Statement.cpp:135] SQLite::Statement::Run INSERT INTO Changes VALUES(NULL, ?, ?, ?, ?)
T0515 14:20:12.527632 ServerIndex.cpp:181] Change related to resource cb042892-c68e4b32-c1d4c890-35e43dfd-09cd9344 of type Study: StableStudy
T0515 14:20:12.528041 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM Resources WHERE internalId=?
T0515 14:20:12.528341 Statement.cpp:135] SQLite::Statement::Run INSERT INTO Changes VALUES(NULL, ?, ?, ?, ?)
T0515 14:20:12.532661 ServerIndex.cpp:181] Change related to resource fdabc3bc-5434e797-74d06b99-fa9248f8-9c3d9ca6 of type Patient: StablePatient
T0515 14:20:12.532950 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:20:12.533247 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:20:12.533498 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:20:12.533760 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:20:12.534160 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:20:12.534499 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.534778 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.535094 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.535300 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.535482 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.535707 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.535904 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.536105 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.536294 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.536488 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.536683 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.536857 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.537041 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.537222 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.537460 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.537641 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.537844 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.538052 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:20:12.538275 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:20:12.538458 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:20:12.538644 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.538908 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:20:12.539160 Statement.cpp:144] SQLite::Statement::Step SELECT parentId FROM Resources WHERE internalId=?
T0515 14:20:12.539376 Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?
T0515 14:20:12.539569 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:20:12.539879 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:20:12.540172 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:20:12.540380 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.540546 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.540703 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.540894 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.541033 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.541202 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.541372 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.541537 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.541707 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.541867 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.542035 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.542208 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.542400 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.542651 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.542811 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.543034 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.543206 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:20:12.543415 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:20:12.543613 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:20:12.543753 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.543999 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:20:12.544234 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:20:12.544525 Statement.cpp:144] SQLite::Statement::Step SELECT a.publicId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?
T0515 14:20:12.544786 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.544962 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.545123 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.545285 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.545454 Statement.cpp:144] SQLite::Statement::Step SELECT * FROM MainDicomTags WHERE id=?
T0515 14:20:12.545633 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.545848 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.546014 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?
T0515 14:20:12.546212 Statement.cpp:144] SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?
T0515 14:20:12.546473 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:20:12.546680 Statement.cpp:144] SQLite::Statement::Step SELECT type FROM Metadata WHERE id=?
T0515 14:20:12.546866 Statement.cpp:144] SQLite::Statement::Step SELECT value FROM Metadata WHERE id=? AND type=?

While querying for such a long time, my DICOM clients seem to time out.

Any idea? Thanks!

According to your log, the WADO-URI command starts at:

I0515 14:19:39.470968 MongooseServer.cpp:723] GET /wado

And ends at:

T0515 14:19:39.776787 HttpOutput.cpp:368] Compressing a HTTP answer using gzip

This means that only 300ms were spent within Orthanc… certainly not enough to trigger a timeout in the client. I am thus really unable to help you.

You should try to see what is the exact HTTP request Horos/OsiriX is issuing (e.g. with Wireshark).

It does time out before doing the WADO command.

Unfortunately, I cannot give any additional insight by myself without an Apple computer.

Could someone else help? Or maybe contact the Horos team?

When I limit "LimitFindInstances" to e.g. 10, it aborts when trying to retrieve files. For me it seems that every single instance in my storage is first queried by sql before doing the WADO command, and when I have lots of patients it just takes too long time... (SQLite::Statement::Step SELECT internalId FROM Resources WHERE resourceType=?;Statement.cpp:144] SQLite::Statement::Step SELECT publicId FROM Resources WHERE internalId=?;SQLite::Statement::Step SELECT internalId, resourceType FROM Resources WHERE publicId=?;Statement.cpp:144] SQLite::Statement::Step SELECT uuid, uncompressedSize, compressionType, compressedSize, uncompressedMD5, compressedMD5 FROM AttachedFiles WHERE id=? AND fileType=?)

Is this expected behavior? I don't think its specific to Apple Software??

Hello,

Indeed, this is most probably a performance issue on the DICOMweb plugin side… but I need to be able to reproduce it without an access to an Apple computer.

Please could you try and reproduce it with cross-platform tools, i.e. either with curl or Python (calling the REST API) or with the DCMTK command-line tools (findscu, movescu, storescu), then post the problematic commands here?

Sébastien-

I know this is not particularly helpful for debugging, but I want to mention that I see the same problem with Mac OS X and Horos.

I get a timeout from Horos while the SQL database is queried, with the following lines repeated many times:

T0527 11:07:43.973690 Statement.cpp:144] SQLite::Statement::Step SELECT a.internalId FROM Resources AS a, Resources AS b WHERE a.parentId = b.internalId AND b.internalId = ?