Hello,

I am not able to perform a query and retrieve from the Orthanc PACS.
I think “DicomModalities” does no function.

It is configured like this:

“DicomModalities” : {

“IMAGINGSYSTEM” : [ “IMAGINGSYSTEM”, “192.168.0.100”, portnumber of imaging system]

},

I am able to store to Orthanc PACS, but when i set “DicomAlwaysAllowStore” : true, to false, It fails. My conclusion is that the “DicomModalities” is not seen?

Regards, Erik

It has never not worked for me. Are you sure it isn’t commented out? Can you provide a full copy of config?

Hi,

Here is the conig file:

{
/**

  • General configuration of Orthanc
    **/

// The logical name of this instance of Orthanc. This one is
// displayed in Orthanc Explorer and at the URI “/system”.
“Name” : “MyOrthanc”,

// Path to the directory that holds the heavyweight files (i.e. the
// raw DICOM instances). Backslashes must be either escaped by
// doubling them, or replaced by forward slashes “/”.
“StorageDirectory” : “C:\Orthanc”,

// Path to the directory that holds the SQLite index (if unset, the
// value of StorageDirectory is used). This index could be stored on
// a RAM-drive or a SSD device for performance reasons.
“IndexDirectory” : “C:\Orthanc”,

// Enable the transparent compression of the DICOM instances
“StorageCompression” : false,

// Maximum size of the storage in MB (a value of “0” indicates no
// limit on the storage size)
“MaximumStorageSize” : 0,

// Maximum number of patients that can be stored at a given time
// in the storage (a value of “0” indicates no limit on the number
// of patients)
“MaximumPatientCount” : 0,

// List of paths to the custom Lua scripts that are to be loaded
// into this instance of Orthanc
“LuaScripts” : [
],

// List of paths to the plugins that are to be loaded into this
// instance of Orthanc (e.g. “./libPluginTest.so” for Linux, or
// “./PluginTest.dll” for Windows). These paths can refer to
// folders, in which case they will be scanned non-recursively to
// find shared libraries. Backslashes must be either escaped by
// doubling them, or replaced by forward slashes “/”.
“Plugins” : [ “…/Plugins/” ],

// Maximum number of processing jobs that are simultaneously running
// at any given time. A value of “0” indicates to use all the
// available CPU logical cores. To emulate Orthanc <= 1.3.2, set
// this value to “1”.
“ConcurrentJobs” : 2,

/**

  • Configuration of the HTTP server
    **/

// Enable the HTTP server. If this parameter is set to “false”,
// Orthanc acts as a pure DICOM server. The REST API and Orthanc
// Explorer will not be available.
“HttpServerEnabled” : true,

// HTTP port for the REST services and for the GUI
“HttpPort” : 8042,

// When the following option is “true”, if an error is encountered
// while calling the REST API, a JSON message describing the error
// is put in the HTTP answer. This feature can be disabled if the
// HTTP client does not properly handles such answers.
“HttpDescribeErrors” : true,

// Enable HTTP compression to improve network bandwidth utilization,
// at the expense of more computations on the server. Orthanc
// supports the “gzip” and “deflate” HTTP encodings.
“HttpCompressionEnabled” : true,

/**

  • Configuration of the DICOM server
    **/

// Enable the DICOM server. If this parameter is set to “false”,
// Orthanc acts as a pure REST server. It will not be possible to
// receive files or to do query/retrieve through the DICOM protocol.
“DicomServerEnabled” : true,

// The DICOM Application Entity Title
“DicomAet” : “ORTHANC”,

// Check whether the called AET corresponds to the AET of Orthanc
// during an incoming DICOM SCU request
“DicomCheckCalledAet” : true,

// The DICOM port
“DicomPort” : 4242,

// The default encoding that is assumed for DICOM files without
// “SpecificCharacterSet” DICOM tag, and that is used when answering
// C-Find requests (including worklists). The allowed values are
// “Ascii”, “Utf8”, “Latin1”, “Latin2”, “Latin3”, “Latin4”,
// “Latin5”, “Cyrillic”, “Windows1251”, “Arabic”, “Greek”, “Hebrew”,
// “Thai”, “Japanese”, and “Chinese”.
“DefaultEncoding” : “Latin1”,

// The transfer syntaxes that are accepted by Orthanc C-Store SCP
“DeflatedTransferSyntaxAccepted” : true,
“JpegTransferSyntaxAccepted” : true,
“Jpeg2000TransferSyntaxAccepted” : true,
“JpegLosslessTransferSyntaxAccepted” : true,
“JpipTransferSyntaxAccepted” : true,
“Mpeg2TransferSyntaxAccepted” : true,
“RleTransferSyntaxAccepted” : true,

// Whether Orthanc accepts to act as C-Store SCP for unknown storage
// SOP classes (aka. “promiscuous mode”)
“UnknownSopClassAccepted” : true,

// Set the timeout (in seconds) after which the DICOM associations
// are closed by the Orthanc SCP (server) if no further DIMSE
// command is received from the SCU (client).
“DicomScpTimeout” : 30,

/**

  • Security-related options for the HTTP server
    **/

// Whether remote hosts can connect to the HTTP server
“RemoteAccessAllowed” : true,

// Whether or not SSL is enabled
“SslEnabled” : false,

// Path to the SSL certificate in the PEM format (meaningful only if
// SSL is enabled)
“SslCertificate” : “certificate.pem”,

// Whether or not the password protection is enabled
“AuthenticationEnabled” : false,

// The list of the registered users. Because Orthanc uses HTTP
// Basic Authentication, the passwords are stored as plain text.
“RegisteredUsers” : {
// “alice” : “alicePassword”
},

/**

  • Network topology
    **/

// The list of the known DICOM modalities
“DicomModalities” : {

“MODALITY” : [ “IENGINE_SCU”, “192.168.0.100”, 29536, ],
“PACS” : [ “PACS”, “192.168.0.106”, 104 ]

},

// Whether to store the DICOM modalities in the Orthanc database
// instead of in this configuration file (new in Orthanc 1.5.0)
“DicomModalitiesInDatabase” : true,

// Whether the Orthanc SCP allows incoming C-Echo requests, even
// from SCU modalities it does not know about (i.e. that are not
// listed in the “DicomModalities” option above). Orthanc 1.3.0
// is the only version to behave as if this argument was set to “false”.
“DicomAlwaysAllowEcho” : true,

// Whether the Orthanc SCP allows incoming C-Store requests, even
// from SCU modalities it does not know about (i.e. that are not
// listed in the “DicomModalities” option above)
“DicomAlwaysAllowStore” : true,

// Whether Orthanc checks the IP/hostname address of the remote
// modality initiating a DICOM connection (as listed in the
// “DicomModalities” option above). If this option is set to
// “false”, Orthanc only checks the AET of the remote modality.
“DicomCheckModalityHost” : true,

// The timeout (in seconds) after which the DICOM associations are
// considered as closed by the Orthanc SCU (client) if the remote
// DICOM SCP (server) does not answer.
“DicomScuTimeout” : 10,

// The list of the known Orthanc peers
“OrthancPeers” : {
/**

/**

  • This is another, more advanced format to define Orthanc
  • peers. It notably allows to specify HTTP headers, a HTTPS
  • client certificate in the PEM format (as in the “–cert” option
  • of curl), or to enable PKCS#11 authentication for smart cards.
    **/
    // “peer” : {
    // “Url” : “http://127.0.0.1:8043/”,
    // “Username” : “alice”,
    // “Password” : “alicePassword”,
    // “HttpHeaders” : { “Token” : “Hello world” },
    // “CertificateFile” : “client.crt”,
    // “CertificateKeyFile” : “client.key”,
    // “CertificateKeyPassword” : “certpass”,
    // “Pkcs11” : false
    // }
    },

// Whether to store the Orthanc peers in the Orthanc database
// instead of in this configuration file (new in Orthanc 1.5.0)
“OrthancPeersInDatabase” : false,

// Parameters of the HTTP proxy to be used by Orthanc. If set to the
// empty string, no HTTP proxy is used. For instance:
// “HttpProxy” : “192.168.0.1:3128”
// “HttpProxy” : “proxyUser:proxyPassword@192.168.0.1:3128”
“HttpProxy” : “”,

// If set to “true”, debug messages from libcurl will be issued
// whenever Orthanc makes an outgoing HTTP request. This is notably
// useful to debug HTTPS-related problems.
“HttpVerbose” : false,

// Set the timeout for HTTP requests issued by Orthanc (in seconds).
“HttpTimeout” : 10,

// Enable the verification of the peers during HTTPS requests. This
// option must be set to “false” if using self-signed certificates.
// Pay attention that setting this option to “false” results in
// security risks!
// Reference: http://curl.haxx.se/docs/sslcerts.html
“HttpsVerifyPeers” : true,

// Path to the CA (certification authority) certificates to validate
// peers in HTTPS requests. From curl documentation (“–cacert”
// option): “Tells curl to use the specified certificate file to
// verify the peers. The file may contain multiple CA
// certificates. The certificate(s) must be in PEM format.” On
// Debian-based systems, this option can be set to
// “/etc/ssl/certs/ca-certificates.crt”
“HttpsCACertificates” : “”,

/**

  • Advanced options
    **/

// Dictionary of symbolic names for the user-defined metadata. Each
// entry must map an unique string to an unique number between 1024
// and 65535. Reserved values:
// - The Orthanc whole-slide imaging plugin uses metadata 4200
“UserMetadata” : {
// “Sample” : 1024
},

// Dictionary of symbolic names for the user-defined types of
// attached files. Each entry must map an unique string to an unique
// number between 1024 and 65535. Optionally, a second argument can
// provided to specify a MIME content type for the attachment.
“UserContentType” : {
// “sample” : 1024
// “sample2” : [ 1025, “application/pdf” ]
},

// Number of seconds without receiving any instance before a
// patient, a study or a series is considered as stable.
“StableAge” : 60,

// By default, Orthanc compares AET (Application Entity Titles) in a
// case-insensitive way. Setting this option to “true” will enable
// case-sensitive matching.
“StrictAetComparison” : false,

// When the following option is “true”, the MD5 of the DICOM files
// will be computed and stored in the Orthanc database. This
// information can be used to detect disk corruption, at the price
// of a small performance overhead.
“StoreMD5ForAttachments” : true,

// The maximum number of results for a single C-FIND request at the
// Patient, Study or Series level. Setting this option to “0” means
// no limit.
“LimitFindResults” : 0,

// The maximum number of results for a single C-FIND request at the
// Instance level. Setting this option to “0” means no limit.
“LimitFindInstances” : 0,

// The maximum number of active jobs in the Orthanc scheduler. When
// this limit is reached, the addition of new jobs is blocked until
// some job finishes.
“LimitJobs” : 10,

// If this option is set to “true” (default behavior until Orthanc
// 1.3.2), Orthanc will log the resources that are exported to other
// DICOM modalities or Orthanc peers, inside the URI
// “/exports”. Setting this option to “false” is useful to prevent
// the index to grow indefinitely in auto-routing tasks (this is the
// default behavior since Orthanc 1.4.0).
“LogExportedResources” : false,

// Enable or disable HTTP Keep-Alive (persistent HTTP
// connections). Setting this option to “true” prevents Orthanc
// issue #32 (“HttpServer does not support multiple HTTP requests in
// the same TCP stream”), but can possibly slow down HTTP clients
// that do not support persistent connections. The default behavior
// used to be “false” in Orthanc <= 1.5.1. Setting this option to
// “false” is also recommended if Orthanc is compiled against
// Mongoose.
“KeepAlive” : true,

// Enable or disable Nagle’s algorithm. Only taken into
// consideration if Orthanc is compiled to use CivetWeb. Experiments
// show that best performance can be obtained by setting both
// “KeepAlive” and “TcpNoDelay” to “true”. Beware however of
// caveats: https://eklitzke.org/the-caveats-of-tcp-nodelay
“TcpNoDelay” : true,

// If this option is set to “false”, Orthanc will run in index-only
// mode. The DICOM files will not be stored on the drive. Note that
// this option might prevent the upgrade to newer versions of Orthanc.
“StoreDicom” : true,

// DICOM associations initiated by Lua scripts are kept open as long
// as new DICOM commands are issued. This option sets the number of
// seconds of inactivity to wait before automatically closing a
// DICOM association used by Lua. If set to 0, the connection is
// closed immediately.
“DicomAssociationCloseDelay” : 5,

// Maximum number of query/retrieve DICOM requests that are
// maintained by Orthanc. The least recently used requests get
// deleted as new requests are issued.
“QueryRetrieveSize” : 10,

// When handling a C-Find SCP request, setting this flag to “true”
// will enable case-sensitive match for PN value representation
// (such as PatientName). By default, the search is
// case-insensitive, which does not follow the DICOM standard.
“CaseSensitivePN” : false,

// Configure PKCS#11 to use hardware security modules (HSM) and
// smart cards when carrying on HTTPS client authentication.
/**
“Pkcs11” : {
“Module” : “/usr/local/lib/libbeidpkcs11.so”,
“Module” : “C:/Windows/System32/beidpkcs11.dll”,
“Pin” : “1234”,
“Verbose” : true
}
**/

// If set to “false”, Orthanc will not load its default dictionary
// of private tags. This might be necessary if you cannot import a
// DICOM file encoded using the Implicit VR Endian transfer syntax,
// and containing private tags: Such an import error might stem from
// a bad dictionary. You can still list your private tags of
// interest in the “Dictionary” configuration option below.
“LoadPrivateDictionary” : true,

// Locale to be used by Orthanc. Currently, only used if comparing
// strings in a case-insensitive way. It should be safe to keep this
// value undefined, which lets Orthanc autodetect the suitable locale.
// “Locale” : “en_US.UTF-8”,

// Register a new tag in the dictionary of DICOM tags that are known
// to Orthanc. Each line must contain the tag (formatted as 2
// hexadecimal numbers), the value representation (2 upcase
// characters), a nickname for the tag, possibly the minimum
// multiplicity (> 0 with defaults to 1), possibly the maximum
// multiplicity (0 means arbitrary multiplicity, defaults to 1), and
// possibly the Private Creator (for private tags).
“Dictionary” : {
// “0014,1020” : [ “DA”, “ValidationExpiryDate”, 1, 1 ]
// “00e1,10c2” : [ “UI”, “PET-CT Multi Modality Name”, 1, 1, “ELSCINT1” ]
// “7053,1003” : [ “ST”, “Original Image Filename”, 1, 1, “Philips PET Private Group” ]
// “2001,5f” : [ “SQ”, “StackSequence”, 1, 1, “Philips Imaging DD 001” ]
},

// Whether to run DICOM C-Move operations synchronously. If set to
// “false” (asynchronous mode), each incoming C-Move request results
// in the creation of a new background job. Up to Orthanc 1.3.2, the
// implicit behavior was to use synchronous C-Move (“true”). Between
// Orthanc 1.4.0 and 1.4.2, the default behavior was set to
// asynchronous C-Move (“false”). Since Orthanc 1.5.0, the default
// behavior is back to synchronous C-Move (“true”, which ensures
// backward compatibility with Orthanc <= 1.3.2).
“SynchronousCMove” : true,

// Maximum number of completed jobs that are kept in memory. A
// processing job is considered as complete once it is tagged as
// “Success” or “Failure”. Since Orthanc 1.5.0, a value of “0”
// indicates to keep no job in memory (i.e. jobs are removed from
// the history as soon as they are completed), which prevents the
// use of some features of Orthanc (typically, synchronous mode in
// REST API) and should be avoided for non-developers.
“JobsHistorySize” : 10,

// Whether to save the jobs into the Orthanc database. If this
// option is set to “true”, the pending/running/completed jobs are
// automatically reloaded from the database if Orthanc is stopped
// then restarted (except if the “–no-jobs” command-line argument
// is specified). This option should be set to “false” if multiple
// Orthanc servers are using the same database (e.g. if PostgreSQL
// or MariaDB/MySQL is used).
“SaveJobs” : true,

// Specifies how Orthanc reacts when it receives a DICOM instance
// whose SOPInstanceUID is already stored. If set to “true”, the new
// instance replaces the old one. If set to “false”, the new
// instance is discarded and the old one is kept. Up to Orthanc
// 1.4.1, the implicit behavior corresponded to “false”.
“OverwriteInstances” : false,

// Maximum number of ZIP/media archives that are maintained by
// Orthanc, as a response to the asynchronous creation of archives.
// The least recently used archives get deleted as new archives are
// generated. This option was introduced in Orthanc 1.5.0, and has
// no effect on the synchronous generation of archives.
“MediaArchiveSize” : 1,

// Performance setting to specify how Orthanc accesses the storage
// area during C-FIND. Three modes are available: (1) “Always”
// allows Orthanc to read the storage area as soon as it needs an
// information that is not present in its database (slowest mode),
// (2) “Never” prevents Orthanc from accessing the storage area, and
// makes it uses exclusively its database (fastest mode), and (3)
// “Answers” allows Orthanc to read the storage area to generate its
// answers, but not to filter the DICOM resources (balance between
// the two modes). By default, the mode is “Always”, which
// corresponds to the behavior of Orthanc <= 1.5.0.
“StorageAccessOnFind” : “Always”
}

Try getting rid of that extra comma after 29536:

“MODALITY” : [ “IENGINE_SCU”, “192.168.0.100”, 29536**,** ],

Hi Bryan, thanks, i took away the comma and set "Whether to store the DICOM modalities in the Orthanc database
// instead of in this configuration file (new in Orthanc 1.5.0)
“DicomModalitiesInDatabase” : false," from false to true. Now the Dicom nodes become visible in the GUI and are selectable. So things get better!

Thanks for your time.
Regards, Erik