MRI reference lines Philips Intera SW ver 11.8

First off,
Much thanks to you for your wonderful contributions to the medical
industry. You have for sure saved many lives This is greatly appreciated !
The Orthanc PACS is very impressive !

My one quirk I am trying hard to resolve seems to be unique to the Philips
Intera MRI (likely all versions of PMS sw, but using 11.8 here
Problem:
No Reference lines . works on Efilm 1.5 and 2.1.2 versions however*
On those versions working, a DICOM dump shows
0008,1100 referenced sequence SQ followed by-
0008,1145 referenced sequence image
On the non-working versions + Orthanc we get a different dump-
0008,1100 referenced sequence SQ followed by-

0008,9092 referenced sequence evidence.
and the data appears on a separate (extra tree branch).

In all cases get successful transfer, no error messages, but the lines are
missing when vied on the0008,9092 ones , newer software ?
I can get dicom dumps and photos if it would be needed.
Thanks so much,
Dave Adams

Hi Dave,

Yes, we would definitely need sample anonymized DICOM files in order to investigate.

It’s also not clear to me in which viewer the reference lines are missing ?
Note that the Osimis Viewer is not maintained anymore since it has been replaced by the StoneViewer (documentation).

Best regards,

Alain

ok i will create 3 sets exported from the Intera MR.
1) Export to Efilm v 2.1.2, then relayed to Orthanc. Works, shows the
reference lines and keeps them.

2) Export directly to Orthanc from Intera - Loses reference lines,
3) Export directly to Efilm v4.3 (same results all version above Efilm
v2.1.2 , Then relayed to Orthanc- Loses reference lines.
Also I will provide dicom dumps (photos if I can't get a text file.
Thank You.