Hi Alain,
I’ve done a little more digging and it is indeed a very common and standard PACS system - It is an Agfa Impax 6.2.x\6.3.x system,
Their Implementation Class UID: 1.2.124.113532.3510.50
Their Implementation Version Name: AGFAMAY2003
I have reviewed the Conformance Statement and it does indeed make mention of this status code. The document can be found at:
https://www.agfahealthcare.com/global/en/he/library/libraryopen?ID=13773625
I refer you to page 37 of 77, table 2.2-34.
It would be very helpful to have a generic method of ignoring or mapping certain status codes. Unfortunately this is not the only instance where I have seen this; some other vendors do the same. By adding a manufacturer specific exception, you may end up chasing your tail with other vendors too.
Kind regards
Stefan