Hello,
I’m getting:
E0705 12:10:16.143091 CommandDispatcher.cpp:877] DIMSE failure (aborting association): DIMSE No data available (timeout in non-blocking mode)
While trying to retrieve big studies (lots of images).
I do not get this error when retrieving smaller studies.
I think I get this error because it take a long time before the first image gets send to Orthanc (the sending server first has to retrieve the complete study from nearline storage).
I believe this can be fixed by increasing the DIMSE C-move timeout.
I cannot find a configuration item to set this timeout.
Can you make this timeout configurable?
Thanks in advance.
Regards,
Robert
Dear Robert,
I have just added two new options to fine-tune the DICOM timeouts:
https://bitbucket.org/sjodogne/orthanc/commits/fabf7820d1f194149f15f0d40123dda498db5e08
// Set the timeout (in seconds) after which the DICOM associations
// are closed by the Orthanc SCP (server) if no further DIMSE
// command is received from the SCU (client).
“DicomScpTimeout” : 30,
// The timeout (in seconds) after which the DICOM associations are
// considered as closed by the Orthanc SCU (client) if the remote
// DICOM SCP (server) does not answer.
“DicomScuTimeout” : 10,
In your case, you would need to increase “DicomScpTimeout”.
Please could you tell me whether these options solve your issue?
Regards,
Sébastien-